Old 08-20-2024, 08:41 AM
Chris K Chris K is offline
VideoKarma Member
 
Join Date: Jun 2022
Posts: 1,534
Quote:
Originally Posted by Penthode View Post
I count four interconnection hookup cables

Blanking
Sweep
Sweep Inhibit
Aux out of the Sweeper tthe counter in rear.

In addition you need the geberator RF out to apply to the circuit under test and a cable, typically scope probe to it's vertical input.

Have you verified that the HP8601a produces a CW output which frequency can be measured on the HP8600a?

Refer to the Operator check on pages 21 and 22 of the BAMA HP8601a manual pdf.
So I have a few questions about this. Some of them will sound like I have no idea what I am doing and in several cases, that's absolutely true! I would like to start with the operator checks on the pages you specified.

Regarding the test setup below...



The connections shown are the only connections needed for this part of the evaluation? Blanking and sweep etc between the 8600 and the 8601 on the rear of both instruments do not need to be connected correct?

What does CW mean? Center wavelength?

RF Detector? What do I use for this? In an analysis setup, it would be the device under test but shouldn't there be a standard for it in a check setup?

For the scope inputs, the sweep output goes to the horizontal (X) input and the output of the RF Detector (whatever that is!) goes to the "Y" input? My scope needs to be in X-Y mode correct?

There...I've put my ignorance out there for all to see! Sometimes, the terms you guys throw around go right over my head! More idiotic questions to come!
Reply With Quote
  #2  
Old 08-20-2024, 07:46 PM
Penthode's Avatar
Penthode Penthode is offline
VideoKarma Member
 
Join Date: Jun 2009
Location: Kitchener/Waterloo Ontario Canada
Posts: 1,463
Quote:
Originally Posted by Chris K View Post
So I have a few questions about this. Some of them will sound like I have no idea what I am doing and in several cases, that's absolutely true! I would like to start with the operator checks on the pages you specified.

Regarding the test setup below...



The connections shown are the only connections needed for this part of the evaluation? Blanking and sweep etc between the 8600 and the 8601 on the rear of both instruments do not need to be connected correct?

What does CW mean? Center wavelength?

RF Detector? What do I use for this? In an analysis setup, it would be the device under test but shouldn't there be a standard for it in a check setup?

For the scope inputs, the sweep output goes to the horizontal (X) input and the output of the RF Detector (whatever that is!) goes to the "Y" input? My scope needs to be in X-Y mode correct?

There...I've put my ignorance out there for all to see! Sometimes, the terms you guys throw around go right over my head! More idiotic questions to come!
Yes that is correct on both counts. we want to verify the thing is working.

The blanking and sweep inhibit are to display the markers on the scope. Instead of creating a beat blip marker where you combine the marker generator with the sweep signal at the input or output, this system introduces a marker by momentarily halting the sweep at the frequency point to make it appear brighter.

For the RF detector, I used a simple diode capacitor resistor network like the one attached. The diode i used was the ubiquitous 1N4148. It essentially rectifies and filters to remove the RF carrier and rf envelope to register on the scope. Try bypassing it and providing the y input of the scope has sufficient bandwidth (eg a 100MHz scope) bypassing it you will see the rf carrier sweep.

I have attached the use of this equipment to obtain my IF response with markers which went through the RF detector of the television set as yours will and a sweep of the luma reponse of the set minus an RF detector. This will demonstrate the presence and absence of the RF detector.

Last edited by Penthode; 09-22-2024 at 12:37 PM.
Reply With Quote
  #3  
Old 08-20-2024, 08:20 PM
Chris K Chris K is offline
VideoKarma Member
 
Join Date: Jun 2022
Posts: 1,534
Quote:
Originally Posted by Penthode View Post
Yes that is correct on both counts. we want to verify the thing is working.

The blanking and sweep inhibit are to display the markers on the scope. Instead of creating a beat blip marker where you combine the marker generator with the sweep signal at the input or output, this system introduces a marker by momentarily halting the sweep at the frequency point to make it appear brighter.

For the RF detector, I used a simple diode capacitor resistor network like the one attached. The diode i used was the ubiquitous 1N4148. It essentially rectifies and filters to remove the RF carrier and rf envelope to register on the scope. Try bypassing it and providing the y input of the scope has sufficient bandwidth (eg a 100MHz scope) bypassing it you will see the rf carrier sweep.
I'll make the detector. Thank you for the design! There may be issues with the 8601. I don't think I'm getting RF output. Output level doesn't change. I'll need to break into it and see what's going on!
Reply With Quote
  #4  
Old 08-20-2024, 09:05 PM
Penthode's Avatar
Penthode Penthode is offline
VideoKarma Member
 
Join Date: Jun 2009
Location: Kitchener/Waterloo Ontario Canada
Posts: 1,463
Quote:
Originally Posted by Chris K View Post
I'll make the detector. Thank you for the design! There may be issues with the 8601. I don't think I'm getting RF output. Output level doesn't change. I'll need to break into it and see what's going on!
Without the HP8600a, the HP8601a should act as a standard signal generator. If you have a 100MHz scope, simply connect the probe to the RF output and set it to say 10MHz. Switch it to CW mode and you can ignore the sweep mode switches for a moment. Just crank up the output and set the TIME/DIV on the scope to 0.1uS per division and see if you can see a high frequency sinusoid on the scope.
Reply With Quote
  #5  
Old 08-20-2024, 09:21 PM
Penthode's Avatar
Penthode Penthode is offline
VideoKarma Member
 
Join Date: Jun 2009
Location: Kitchener/Waterloo Ontario Canada
Posts: 1,463
Here is the setting of my HP8601a generating a 10MHz CW carrier and the scope set to 0.1us per division and scope set to 0.1volts/div with a 10x probe = 1.0 volts per division.
Attached Images
File Type: jpg HP8601a_level.jpg (117.2 KB, 11 views)
File Type: jpg 8601a_scope.jpg (120.7 KB, 12 views)
Reply With Quote
Audiokarma
  #6  
Old 08-20-2024, 11:20 PM
Penthode's Avatar
Penthode Penthode is offline
VideoKarma Member
 
Join Date: Jun 2009
Location: Kitchener/Waterloo Ontario Canada
Posts: 1,463
Ah! in the test circuit I forgot the most important element: the diode!

Here is the revised drawing.
Attached Images
File Type: jpg Test circuit_a.jpg (21.0 KB, 10 views)
Reply With Quote
  #7  
Old 08-21-2024, 09:41 AM
Chris K Chris K is offline
VideoKarma Member
 
Join Date: Jun 2022
Posts: 1,534
I think this is just noise. Output level looks dead. There's a guy on eBay selling these restored/calibrated and certified accurate. Maybe I'll look into those units if I can't get this one working. My ongoing luck with this endeavor and obtaining TVs from this era that require a recap and a few tweaks to get a good picture continues. Kinda ridiculous at this point. I feel cursed!





Reply With Quote
Reply


Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off

Forum Jump


All times are GMT -5. The time now is 12:25 AM.



Powered by vBulletin® Version 3.8.4
Copyright ©2000 - 2026, Jelsoft Enterprises Ltd.
©Copyright 2012 VideoKarma.org, All rights reserved.